It's an old Hitachi S-450, made around 1980 or so. Came with all sorts of goodies which even fit the darn thing didn't work (Which they said it did not) it would be worth it. Came with a Kevex X-ray spectrometer (8" Floppies!), a 50l dewar for LN2, Edwards #12 direct drive pump, and a whole box of spare parts, stubs, etc. They told us they stopped using it about 10 years ago when the board in the X-Ray spec went bad. One of the other guys also mentioned something about a bad tube.
A couple weeks after we got it back to my friends lab we put it back together. Eventually it pumped down and tried to get an image, nothing. I machined a piece of acrylic to make a window for one of the accessory ports on the chamber and coated a stub with phosphor from a old fluorescent lamp. From this we were able to tell that we were getting an e-beam and you could see the raster on the phosphor. After that I got a HV probe from a friend and checked the other high voltage points. We were getting about +500v to the SED detector screen but when I checked the PMT power supply in the cabinet I found it was completely dead. The transformer was warm on it though so I figured it was the output diodes.
Remembering that I had noticed the cable had been spliced going to PMT head I decided to see what that was all about. I pulled the electrical tape off and found the problem. It seems some genius decided it would be too much trouble to remove the lid and disconnect the cable from the console when they went to move it so they just lopped the cable about 2' from the pmt. And then it appears they hired monkeys to solder it back together. The HV line is coax and the splice eventually shorted out and carbon tracked killing the PMT power supply. I ended up just removing the short length and redoing the connections. I hooked up a variable HV supply to the PMT and started to see something before I gave it too much voltage and killed one of the resistors in the PMT divider circuit. I also killed the coax so I replaced that as well. I tried again with a smaller Ultravolt power supply but that did not provide enough current as they used rather low resistance resistors in the PMT divider circuit. Since that was a bust I decided to just wait till the new diodes for the old power supply came in.
Today I installed the diodes and it works. buttoned everything up. Next is to try and figure out what is going on with the X-ray unit. Also need to figure out how to align the aperture. We have been running with it on the retracted position which does kill the DOF.
Here are some pics. The blue one was taken with a tripod in front of the screen and using a slow scan.

IMG_1340 by macona, on Flickr

IMG_1353 by macona, on Flickr

IMG_1350 by macona, on Flickr

546660_10150789638465614_708640613_9649901_1137363641_n by macona, on Flickr